Eğer iki cihaz aynı olmasına rağmen birisi tek grup çalışırken diğeri bütün gruplara bölüyorsa ya yazılım versiyonu farklı ya da cihaz kurcalandığını anlayıp kasıtlı olarak farklı çalışıyor.
Kafamda kalan tek soru işareti snubber neden fail etti? Acaba test ettiğiniz bir cihazın çıkışı snubber ı fazla mı yordu?

❯ Please review the programming manual and then let's try again. ~/Download/IT-M7700 Programming Guide-EN.pdf
Ran 5 shell commands
I've extracted the guide (it was in ~/Downloads) and I'm reading the
source, protection, multi-channel and measure chapters now.
Ran 3 shell commands
The guide clarifies this unit's actual command set (NORMal:MODE,
NORMal:VOLTage:DC, PROTect:..., MEASure:VOLTage:DC?) versus what I
tried earlier. I'll now check the error queue first, then run only
documented, read-only queries one at a time with 100ms gaps, stopping
at the first one that gets no reply.
Ran 1 shell command
Every query worked except the last, PROT:SENS:CHECK?, which got no
reply. I'll read the error queue once to see whether it logged an
error.
Ran 1 shell command
I've read the IT-M7700 programming guide and probed the IT-M7723E
again using only commands it documents. This time it raised no errors:
the queue was empty before and after. The queries I tried earlier were
guesses that aren't in ITECH's command set. On this unit, setpoints
are under NORMal:, protections under PROTect:, and readings are
MEASure:VOLTage:DC? / MEASure:CURRent:DC?.
Current state (read-only queries, nothing changed):
┌────────────────────────────────────────┬──────────────────────────────────┐
│ Setting │ Value │
├────────────────────────────────────────┼──────────────────────────────────┤
│ Mode │ AC │
├────────────────────────────────────────┼──────────────────────────────────┤
│ Output │ OFF │
├────────────────────────────────────────┼──────────────────────────────────┤
│ DC setpoint │ 400.0 V (DC range −400…+400 V) │
├────────────────────────────────────────┼──────────────────────────────────┤
│ Current range │ AUTO │
├────────────────────────────────────────┼──────────────────────────────────┤
│ Voltage rise time │ 1 ms │
├────────────────────────────────────────┼──────────────────────────────────┤
│ Output relay │ opens and closes with the output │
├────────────────────────────────────────┼──────────────────────────────────┤
│ PROT:MAX:CURR:LIM │ 20 A │
├────────────────────────────────────────┼──────────────────────────────────┤
│ RMS current protection (OCP) │ 3.75 A, 50 delay │
├────────────────────────────────────────┼──────────────────────────────────┤
│ Peak current protection │ 50 A │
├────────────────────────────────────────┼──────────────────────────────────┤
│ Over-voltage protection (RMS and peak) │ 440 V │
├────────────────────────────────────────┼──────────────────────────────────┤
│ Over-power protection │ 1500 W │
└────────────────────────────────────────┴──────────────────────────────────┘
PROT:SENS:CHECK? got no reply and logged no error, so I won't use it.
What the guide says that matters here:
- The stored DC setpoint is 400 V. Switching to DC mode and turning
the output on as it stands would put 400 V on the load. The DC
setpoint has to go to 0 V before the switch to DC.
- SYST:REM is needed before any setting change. It locks the front
panel except for Local and On/Off.
- PROT:MAX:CURR:LIM doesn't behave like CC on a lab supply. The guide
calls it the "max current limit" without describing how it acts. If
it does limit, it's a slow limiter, so I'd treat it and OCP only as
safety nets. The current would come from the series resistor.
Proposed setup sequence. Nothing gets sent until you approve it.
1. SYST:REM, then OUTP OFF.
2. NORM:VOLT:DC 0, then NORM:MODE DC. Read back both settings to confirm.
3. Software limits: NORM:VOLT:DC:MAX 160 and NORM:VOLT:DC:MIN 0. With
these in place, a wrong setpoint can't go above 160 V or
negative. They may be saved on the instrument, so you'd change them
back when you return to AC work.
4. Protections: OVP PROT:RMS:VOLT 170, PROT:PEAK:VOLT 175, OCP
PROT:RMS:CURR 0.2, current limit PROT:MAX:CURR:LIM 0.25,
NORM:CURR:RANG LOW. I'll read each one back. If the instrument
rejects a value (error -222, data out of range), I'll stop and
report rather than try another.
5. Per test point: Vs = Vset + 10 V, capped at 152 V. That gives 100
mA through 100 Ω. The 2450 measures the voltage at the load
terminals (2-wire voltmeter, no sensing by the source). I'll log
the source's own MEAS:VOLT:DC? and MEAS:CURR:DC? for
cross-checking.
Hardware you'll need: a 100 Ω resistor rated at least 2 W (it
dissipates 1 W at 100 mA; 5 W gives margin), in series between the
IT-M7723E's + output and the load's + input. Leave the 2450 across the
load terminals as the voltmeter. Disconnect the DP821A and 2461 from
the circuit.
Do you have the 100 Ω resistor? And shall I send steps 1–4 and read
everything back, with the output still OFF? Before any output-on run
I'll add this source to verify.py as its own mode and show you the
changes.
✻ Worked for 1m 42s · done 12:31 PM
Evet ben de merak ediyorumÇok uzun zaman önce oldu o yüzden hiç hatırlamıyorum. Hani şebekeye bağlamış olsam, 220 nF empedansı 14 kOhm yapıyor, akım düşük. Zaten reverse polarity algılayıp çıkışı da keserdi. Belki de bir SMPS şebeke tarafındaki doğrultucu çıkışına bağladım ve doğrudan 325V gitti. Direnç de ilk akım darbesinde aşırı voltajdan koptu.